Jesito wrote:
Hi colleagues!
I am the proud owner of a Gossen Sixon 2. It was a present of my father-in-law, 40 years ago. For reasons that isn't worth to mention, it has been sitting unused in his box until past year, when I rediscovered it.
Brand new after so many years.
It's very nice. Even if it was designed for the old mercury batteries, supports very well the new alkaline ones. It has a regulation screw that allows to adjust the meter according the battery level.
Later on my daugther got interested in photography. I bought at the eBay another light meter for her. This time it was a Sekonic Auto-Lumi 86. Because the smaller size.
After the first tests, we spot a difference of some 3-4 points between the two light meters.
I quickly thought about the selenium cell on the Sekonic and the ageing of the selenium, so I decided to fix it. If not the cell, maybe the resistor that uses to come in series with the meter had changed its value along the years.
Opening the meter was not difficult: Three screws in the bottom side, two of them covered by wax (I recall that practice long ago, to prevent the screws from rust). I had to scratch the wax until the screw head was visible, and unscrew.
In the inside, what was expected: A 1k1 resistor in series with the selenium cell and the meter:
I unsoldered one of the leads of the resistor, and it measured precisely 1K1... The meter seemes to be fine. The selenium cell as well...
So I resoldered and closed it.
Then I picked it with me and this morning I scheduled a working lunch with my daugther. I took both light meters and the Sunny F16 printed table.
As you know, all these meters work in the same way: The cell and galvanometer give the EV value, that you place on a mechanical wheel on top of the lightmeter. Pointing the arrow to the proper EV, it gives two scales with the corresponding speed/aperture pairs.
So first we tried to measure the EV in several different lighting conditions. To my surprise both meters were giving the same EV measures....
So the problem was lying on the scales!. That was evident: putting the same EV we got two measures separated by 4 points.
Which one was telling the truth?
I picked the Sunny F16 table and applied the EV values that we were getting. And they matched perfectly the Sekonic one!!!!
So, now I'm waiting for an opportunity to disassemble the Gossen. But since what I don't do well is estimating the EV, I'm afraid I'll carry the Gossen as a EV meter and I'll use the Sunny 16 table to derive the speed/aperture pairs....
Never trust anything for the fact that it's brand new....
Jes. _________________ Jesito, Moderator
Jesito's backsack:
Zooms Sigma 70-300, Tamron 35-135 and 70-210 short, 70-210 long, 28-70 CF Macro, 35-70, 35-80, Vivitar 70-210 KA, Tamron 70-250.
Fixed Industar-50, , Tamron 24mm, Tamron 135mm, Sands Hunter 135mm, Pancolar 50mm, Volna-3, many Exakta lenses
DSLR SIGMA SD9 & SD14, EOS 5D, Sony A700 and NEXF3, Oly E-330, E-400, E-450, E-1
TLR/6x6/645 YashicaMat, Petri 6x45, Nettar, Franka Solida, Brilliant
SLR Minolta X300, Fuji STX II, Praktica VLC3, Pentax P30t, EXA500, EXA 1A, Spotmatic(2), Chinon CM-4S, Ricoh, Contax, Konica TC-X , Minolta 5000, 7000i, 3Sxi, EOS 500 and CX
Rangefinders Chinon 35EE, Konica C35 auto, Canonet 28, Yashica Lynx, FED-2, Yashica electro 35, Argus C3 & C4, Regula Cita III, Voigtlander Vitoret (many), Welta Welti-I, Kodak Signette 35, Zorki-4, Bessa-R & L, Minolta Weathermatic, olympus XA2
Compact Film Konica C35V, Voigtlander Vitorets, Canon Prima Super 105, Olympus XA2 and XA3
Compact Digital Olympus C-5050, Aiptek Slim 3000, Canon Powershot A540, Nikon 5200, SIGMA DP1s, Polaroid X530, IXUS55, Kodak 6490, Powershot G9 and G10
CSCCanon EOS-M, Samsung NX100 and NX210, Lumix G5, NEX-F3 |
|
| Volver arriba |
|
 |
peterqd
 Registrado: 28 Feb 2007 Mensajes: 7449 Ubicación: near High Wycombe, UK
Expire: 2014-01-04
|
 Publicado: Vie Nov 07, 2008 12:01 am Asunto: |
|
 Publicado: Vie Nov 07, 2008 12:01 am Asunto: |
|
|
peterqd wrote:
Jes, are you sure the numbers on the Gossen scale are actually EV's? On my Weston meter there are large numbers on the scale representing the numbers on the meter, but there is a little window with the EVs given separately. Keeping the main scale pointed at the meter number, the EV changes as you alter the ASA setting.
For example:
ASA200 Meter Number 8 - EV = 10.3
ASA400 Meter Number 8 - EV = 11.3 _________________ Peter - Moderator |
|
| Volver arriba |
|
 |
Sevo
 Registrado: 22 Ago 2008 Mensajes: 1189 Ubicación: Frankfurt, Germany
Expire: 2012-12-03
|
 Publicado: Vie Nov 07, 2008 12:08 am Asunto: |
|
 Publicado: Vie Nov 07, 2008 12:08 am Asunto: |
|
|
Sevo wrote:
And are you sure the meter can be recalibrated to another voltage source? Often, the 1.35V are assumed to be the one stable value in the system, and adjustments can only compensate for sensor or amperemeter drift - if you change the voltage and adjust the sensitivity to match that, the meter becomes non-linear.
Btw. forget about alkalines, these are drifting continuously throughout their discharge - at the very least you have to use silver cells. If the meter remains non-linear when used with a silver cell, you'll have to solder in a diode to drop the voltage to 1.35V. |
|
| Volver arriba |
|
 |
Jesito
 Registrado: 24 Ago 2007 Mensajes: 5745 Ubicación: Olivella, Catalonia, (Spain)
Expire: 2015-01-07
|
 Publicado: Vie Nov 07, 2008 12:12 am Asunto: |
|
 Publicado: Vie Nov 07, 2008 12:12 am Asunto: |
|
|
Jesito wrote:
| peterqd escribió: |
Jes, are you sure the numbers on the Gossen scale are actually EV's? On my Weston meter there are large numbers on the scale representing the numbers on the meter, but there is a little window with the EVs given separately. Keeping the main scale pointed at the meter number, the EV changes as you alter the ASA setting.
For example:
ASA200 Meter Number 8 - EV = 10.3
ASA400 Meter Number 8 - EV = 11.3 |
Hi Peter, thanks for the suggestion, but in the Gossen the EV doens't change even if you change the ASA....
Jes. _________________ Jesito, Moderator
Jesito's backsack:
Zooms Sigma 70-300, Tamron 35-135 and 70-210 short, 70-210 long, 28-70 CF Macro, 35-70, 35-80, Vivitar 70-210 KA, Tamron 70-250.
Fixed Industar-50, , Tamron 24mm, Tamron 135mm, Sands Hunter 135mm, Pancolar 50mm, Volna-3, many Exakta lenses
DSLR SIGMA SD9 & SD14, EOS 5D, Sony A700 and NEXF3, Oly E-330, E-400, E-450, E-1
TLR/6x6/645 YashicaMat, Petri 6x45, Nettar, Franka Solida, Brilliant
SLR Minolta X300, Fuji STX II, Praktica VLC3, Pentax P30t, EXA500, EXA 1A, Spotmatic(2), Chinon CM-4S, Ricoh, Contax, Konica TC-X , Minolta 5000, 7000i, 3Sxi, EOS 500 and CX
Rangefinders Chinon 35EE, Konica C35 auto, Canonet 28, Yashica Lynx, FED-2, Yashica electro 35, Argus C3 & C4, Regula Cita III, Voigtlander Vitoret (many), Welta Welti-I, Kodak Signette 35, Zorki-4, Bessa-R & L, Minolta Weathermatic, olympus XA2
Compact Film Konica C35V, Voigtlander Vitorets, Canon Prima Super 105, Olympus XA2 and XA3
Compact Digital Olympus C-5050, Aiptek Slim 3000, Canon Powershot A540, Nikon 5200, SIGMA DP1s, Polaroid X530, IXUS55, Kodak 6490, Powershot G9 and G10
CSCCanon EOS-M, Samsung NX100 and NX210, Lumix G5, NEX-F3 |
|
| Volver arriba |
|
 |
Jesito
 Registrado: 24 Ago 2007 Mensajes: 5745 Ubicación: Olivella, Catalonia, (Spain)
Expire: 2015-01-07
|
 Publicado: Vie Nov 07, 2008 12:15 am Asunto: |
|
 Publicado: Vie Nov 07, 2008 12:15 am Asunto: |
|
|
Jesito wrote:
| Sevo escribió: |
And are you sure the meter can be recalibrated to another voltage source? Often, the 1.35V are assumed to be the one stable value in the system, and adjustments can only compensate for sensor or amperemeter drift - if you change the voltage and adjust the sensitivity to match that, the meter becomes non-linear.
Btw. forget about alkalines, these are drifting continuously throughout their discharge - at the very least you have to use silver cells. If the meter remains non-linear when used with a silver cell, you'll have to solder in a diode to drop the voltage to 1.35V. |
Hi Sevo, the measuring part works fine. It has a switch on the back to calibrate according the battery. The run of the screw is enough to cover the voltage difference. In fact, the EV measuring works fine. It's a matter of the scales that are not properly synchronized...
Jes. _________________ Jesito, Moderator
Jesito's backsack:
Zooms Sigma 70-300, Tamron 35-135 and 70-210 short, 70-210 long, 28-70 CF Macro, 35-70, 35-80, Vivitar 70-210 KA, Tamron 70-250.
Fixed Industar-50, , Tamron 24mm, Tamron 135mm, Sands Hunter 135mm, Pancolar 50mm, Volna-3, many Exakta lenses
DSLR SIGMA SD9 & SD14, EOS 5D, Sony A700 and NEXF3, Oly E-330, E-400, E-450, E-1
TLR/6x6/645 YashicaMat, Petri 6x45, Nettar, Franka Solida, Brilliant
SLR Minolta X300, Fuji STX II, Praktica VLC3, Pentax P30t, EXA500, EXA 1A, Spotmatic(2), Chinon CM-4S, Ricoh, Contax, Konica TC-X , Minolta 5000, 7000i, 3Sxi, EOS 500 and CX
Rangefinders Chinon 35EE, Konica C35 auto, Canonet 28, Yashica Lynx, FED-2, Yashica electro 35, Argus C3 & C4, Regula Cita III, Voigtlander Vitoret (many), Welta Welti-I, Kodak Signette 35, Zorki-4, Bessa-R & L, Minolta Weathermatic, olympus XA2
Compact Film Konica C35V, Voigtlander Vitorets, Canon Prima Super 105, Olympus XA2 and XA3
Compact Digital Olympus C-5050, Aiptek Slim 3000, Canon Powershot A540, Nikon 5200, SIGMA DP1s, Polaroid X530, IXUS55, Kodak 6490, Powershot G9 and G10
CSCCanon EOS-M, Samsung NX100 and NX210, Lumix G5, NEX-F3 |
|
| Volver arriba |
|
 |
|
|